Post subject: Tektronix 647A scope noise Posted: Nov Thu 23, 2017 3:57 pm |
Joined: Jan Thu 15, 2015 4:56 pm Posts: 116 Location: Winona, Mn.
Top |
Sol
Post subject: Re: Tektronix 647A scope noise Posted: Nov Thu 23, 2017 4:52 pm |
Member |
Odd that it would come from the vertical section, especially if you are hearing the sound with no signal.
My first guesses would be high voltage section, which is the usual culprit in video devices, or it may be something in the blanking circuit. Does the tone change when you change the TIME/CM control?
It's hard to say if it could damage the scope. It may be a component on the edge of going out, but then again it just may be something a little dirty or leaky. Either way I don't see it getting 'better' without an intervention. Maybe you could get in there and spray around with some cold spray to try and better locate the cause? Or, is it all solid state? perhaps you could consult the schematic and pick out a few key tubes to pull, one at a time, and see if that affects the noise.
Top |
Notimetolooz
Post subject: Re: Tektronix 647A scope noise Posted: Nov Fri 24, 2017 1:34 am |
Joined: May Thu 14, 2015 4:15 pm Posts: 739 Location: Dallas, TX
You might want to try a length of plastic hose or tubing held to your ear to localize the sound. It may be coming from below the circuit card.
_________________ Tim It's not the Destination, It's the Journey.
Top |
Mikeinkcmo
Post subject: Re: Tektronix 647A scope noise Posted: Nov Fri 24, 2017 1:05 pm |
Joined: Oct Sun 15, 2006 12:57 pm Posts: 5540 Location: Liberty, Missorui
What a great Gift!!! Thats a 100 meg scope. Tek used power oscillators to generate their CRT High voltages for decades, and if memory serves, their frequency was on the order of a few KHz, so that's perhaps what you're hearing. Typically their power supplies were located at the rear of the scopes. Get a COPY OF THE SERVICE MANUAL here, and go to page 122 to check out the schematic of the supply. If you go to the table of contents, you'll most likely find the theory of operation for the supply/oscillator.
_________________ Bold text & pics are usually links.
Top |
R_Dubb
Post subject: Re: Tektronix 647A scope noise Posted: Nov Sun 26, 2017 4:50 pm |
Well, I was finally able to sit down and play with this thing again. The tiny high frequency sound is there only when the calibrator selector is on a voltage selection. Sound does not change in amplitude within the voltage range. When in off position there is no sound. (?) The calibrating voltages are pretty much right on. The 100 VDC reference is 600 mv short of 100.
When powering up the scope it takes about 30 seconds for the sweep trace to appear at the bottom of the CRT. It then continues to rise and stabilize, to the previous setting, after a couple of minutes of warming up. (?)
There are 3 5642 half diode tubes and 1 7586 Nuvistor. Rest is solid state components. Picture shows the 3 5642's behind a plexiglass cover.
There are 2 miniature flanged base lamps for illuminating the graticule. Both are open. Mouser has them.
My next move is to check the power supply voltages and also see what ripple is out of spec. Service manual specifies a factory calibrating plugin when checking voltages. I don't have that. (?)
Attachments: |
Top |
Sol
Post subject: Re: Tektronix 647A scope noise Posted: Nov Sun 26, 2017 6:30 pm |
Member |
Perhaps there is a leaky cap in the calibrator multivibrator circuit. Sounds like it's somewhere around there.
Top |
Willitwork
Post subject: Re: Tektronix 647A scope noise Posted: Nov Sun 26, 2017 6:52 pm |
Member |
Joined: Jan Thu 24, 2013 4:55 am Posts: 1239 Location: South New Jersey 08077
Disc ceramic capacitors are capable of being an audio transducer, converting audio range electrical signal to physical audio vibrations. I do not know if that effect has any increase with age. You may find the actual sound you hear coming from a disc ceramic.
_________________ Ex: Same job, same building, same phone #; 4 Corp's: RCA, GE, Martin, Lockheed-Martin: 48 years
Top |
flanneltuba
Post subject: Re: Tektronix 647A scope noise Posted: Nov Sun 26, 2017 7:47 pm |
Member |
Joined: Mar Tue 22, 2016 12:51 am Posts: 219 Location: Creswell, OR.
My father used to use a stethoscope to isolate such difficult to pinpoint sounds, as have I n a couple occasions. Granted around high voltage, this technique would need to be exercised with an abundance of caution, though most stethoscopes use rubber or neoprene tubing, so your head wouldn't necessarily become a current path. But still...
As has been essentially said already, if there's a "singing" component or mechanical part in the device, merely applying a bit of pressure with an insulated rod of some sort selectively throughout the suspect region, ought to alter the volume or tone of the sound when you push on the offending part.
Just a though.
- Scott
Top |
MarkD
Post subject: Re: Tektronix 647A scope noise Posted: Dec Thu 07, 2017 4:57 pm |
I had a 647a operational a few yeas ago, and it, too produced a faint high pitched noise. Something went wrong in the sweep, I never got around to trouble shooting it. I think, however that noise is "normal"
Top |
Barry H Bennett
Post subject: Re: Tektronix 647A scope noise Posted: Dec Fri 08, 2017 1:41 pm |
Joined: Nov Wed 30, 2016 7:35 pm Posts: 844 Location: OHIO 43074
If the noise is only present when the calibrator is on, I wouldn't worry about it too much unless you need a "project"  . Quite possibly, that's normal. It could be that the square wave generated by the calibrator is somehow causing some odd effect with the power supply oscillators... or it could be one of the other suggestions noted. Unless you want to get in there with a "sound tracer" (plastic hose, plastic rod for contact) ... it'll be really tough to localize. Doesn't sound harmful unless it's happening all the time. Question .. do you hear it with the covers ON ???
_________________ He who dies with the most sheer tonnage of ancient test gear, wins!
Top |
ac
Post subject: Re: Tektronix 647A scope noise Posted: Dec Thu 14, 2017 1:57 am |
The calibrator in my 647A also makes that noise. It's normal, and comes from the 4KHz crystal vibrating. Most crystals are well above the audible range, so you normally don't notice this.
The only known major problem with this scope is that the HV transformer can become lossey and go into thermal runaway. When this happens, the fuse for the HV supply will blow after about 30-60 minutes. I have a 647 with this problem, but so far my 647A has been fine. The only permanent solution is a new HV transformer (or to rewind the old one).
Top |
MM
Post subject: Re: Tektronix 647A scope noise Posted: Dec Thu 14, 2017 1:01 pm |
Joined: Nov Wed 15, 2006 7:56 am Posts: 1105 Location: germany
this plastik boxes for HV, also in 547 and 556, are a thermal isolator. Good help is to insert a little CPU fan there
greetings Martin
Top |
R_Dubb
Post subject: Re: Tektronix 647A scope noise Posted: Dec Sun 17, 2017 6:36 pm |
AC, I think I have that HV transformer problem. Awhile back I was playing with it and after about an hour the sweep disappeared. The next day I traced it to F820 (1A fuse) which was blown. I replaced the fuse and it came back to life. This time I payed more attention to the clock. I remembered, the first time around, that I had to adjust the brightness up a couple of times before the fuse blew. I noticed the trace was starting to dim, the second time, at close to an hour of operating time. I turned it off for about 15 minutes then restarted it. Trace was there but had to reduce the intensity. (too bright) So the windings on the HV transformer are shorting causing the the fuse to blow?
Top |