Post subject: now its junk Posted: Mar Sat 31, 2012 11:44 pm |
|
Joined: Jan Thu 01, 1970 1:00 am Posts: 892 Location: Sodus,NY USA
|
|
got given a Acturus "BLUE" globe 127 marked Detector tube today.put it in my tube tester to check it out and something happened with the tester and it suddenly sent a shot of voltage through to the tube and made it go real bright and before I could shut it off the tube made a flash and I guess is now junk.guess now it is nothing but something for display.just my luck.the tester is the only working one that I have that tests 4 and 5 prong tubes.guess that is junk now too.
_________________ Visit us on the web at: sunshineradiomuseum.org W2HJC
|
|
| Top |
|
 |
|
Johnnysan
|
Post subject: Re: now its junk Posted: Apr Sun 01, 2012 12:20 am |
|
Joined: Jan Thu 01, 1970 1:00 am Posts: 11441 Location: Albuquerque, NM 87123
|
|
If it burned out a filament it must have been a major malfunction. Even a shorted line rheostat shouldn't have placed that much voltage on a filament. Are you certain the filament didn't sag and touch the cathode? If so, the tube was a goner anyway.
|
|
| Top |
|
 |
|
hcheetham
|
Post subject: Re: now its junk Posted: Apr Sun 01, 2012 12:25 am |
|
Joined: Jan Thu 01, 1970 1:00 am Posts: 892 Location: Sodus,NY USA
|
|
not sure what did happen.it happended when I was adjusting the line on the meter before starting the test.
_________________ Visit us on the web at: sunshineradiomuseum.org W2HJC
|
|
| Top |
|
 |
|
Tubenut
|
Post subject: Re: now its junk Posted: Apr Sun 01, 2012 2:02 pm |
|
Joined: Oct Sun 11, 2009 10:06 am Posts: 1441 Location: British Columbia
|
|
You Know, you never adjust the tube tester with any tube in the socket. All the adjustments are done first, then the tube gets plugged in.
|
|
| Top |
|
 |
|
Peter Bertini
|
Post subject: Re: now its junk Posted: Apr Mon 02, 2012 1:12 am |
|
| Moderator |
 |
Joined: Jan Thu 01, 1970 1:00 am Posts: 12217 Location: Somers, CT
|
|
I agree with Johnnysan's observations, 100 percent. Adjusting the line calibration would hardly account for this sort of failure. Unless the filament voltage setting was grossly mis-set, there is no logical explanation for the tube failure.
Pete
_________________ A long journey always begins with the words, "I think I know a shortcut."
|
|
| Top |
|
 |
|
wazz
|
Post subject: Re: now its junk Posted: Apr Mon 02, 2012 1:15 am |
|
| Member |
 |
Joined: Jun Wed 08, 2011 2:33 am Posts: 2059 Location: Ohio
|
|
Tube must have had some sort of internal short. If you bought it as a working tube online, you might see what you can wrangle out of the seller. Although usually the internal shorts test is the first thing to do.
|
|
| Top |
|
 |
|
Chas
|
Post subject: Re: now its junk Posted: Apr Mon 02, 2012 1:36 am |
|
| Member |
 |
Joined: Jan Thu 01, 1970 1:00 am Posts: 5259 Location: S. Dartmouth MA 02748-1225 USA
|
|
I suggest that a dummy 5-pin plug set up with a 2.5 volt lamp as a filament load. Then attempt to duplicate the disaster settings and try and roast the lamp. All the while paying strict attention to what control is being manipulated.
YMMV
Chas
_________________ "Don't find fault, find a remedy"
(Ancient Chinese cookie fortune)
|
|
| Top |
|
 |
|
A A Kent
|
Post subject: Re: now its junk Posted: Apr Mon 02, 2012 3:32 pm |
|
Joined: Feb Wed 25, 2009 4:06 pm Posts: 796 Location: Morristown, N.J.
|
|
It's an easy mistake to make for even the most experienced tech/restorer. I always try to remember to turn the filament voltage down all the way when I finish testing tubes and to check for it when I begin to do more tube tests.
Blue 27's are rather common so it's not a great loss. Had it been a blue audio tube, you would need to take away my belt and shoe laces.
Pete
|
|
| Top |
|
 |
|
Chris108
|
Post subject: Re: now its junk Posted: Apr Tue 03, 2012 12:22 am |
|
Joined: Jun Fri 19, 2009 6:34 pm Posts: 2673 Location: Long Island
|
Quote: You Know, you never adjust the tube tester with any tube in the socket. All the adjustments are done first, then the tube gets plugged in. So then how are you supposed to adjust the line voltage control? The manuals for every tube tester I've ever owned tell you to bring the needle to the mark with the tube in, so you are adjusting for the heater load. Basic emission testers often have voltages applied at all times, so no switches or controls should be changed on those while a tube is in a socket. But better quality testers do not apply voltages until the "Test" button is pressed, so it is perfectly safe to change settings as long as you are not holding the button down. Of course, one had better make sure the heater voltage is properly set before plugging a tube in, but that goes for any kind of tube tester. The 27 tube has a 2.5-volt, 1.75-amp heater. This is a strain for the transformers on many tube testers. Consequently, it is doubtful that the line voltage control on the OP's tester would have had enough range to cause the tube to burn out. It is far, far more likely that the filament was brittle from age and handling, and it simply broke. Happens to everybody sooner or later.
_________________ "Hell, there are no rules here--we're trying to accomplish something!"
Thomas A. Edison
|
|
| Top |
|
 |
|
Tubenut
|
Post subject: Re: now its junk Posted: Apr Tue 03, 2012 3:30 am |
|
Joined: Oct Sun 11, 2009 10:06 am Posts: 1441 Location: British Columbia
|
Chris108 wrote: Quote: You Know, you never adjust the tube tester with any tube in the socket. All the adjustments are done first, then the tube gets plugged in. So then how are you supposed to adjust the line voltage control? I have 5 tube testers, never have I had to adjust the line Voltage control with a tube in. A number 27 tube's filament draw is moderate. For the reason above, I think this is a good practice. Especially if it's a special tube. Many old tube testers have a step control for the line adjust, IE: Heathkit and so on...... There is no need to be "jolting" the tube, when setting this control prior to the test is perfectly adequate.
|
|
| Top |
|
 |
|
Unusualdesigner
|
Post subject: Re: now its junk Posted: Apr Tue 03, 2012 4:00 am |
|
| Member |
 |
Joined: May Sun 08, 2011 10:45 pm Posts: 2916 Location: Southern Calif
|
Chris108 wrote: Quote: So then how are you supposed to adjust the line voltage control? The manuals for every tube tester I've ever owned tell you to bring the needle to the mark with the tube in, so you are adjusting for the heater load. Noted but I'd recommend doing the following: Set the load control and all the other settings FIRST, then double-check to make sure they are right, then plug tube in, let it warm up and then re-check the load setting before doing the actual test. Never had a catastrophic failure doing that.
_________________ It's easier to make a steam powered lobster trap than a lobster powered steam trap
|
|
| Top |
|
 |
|
GordonW
|
Post subject: Re: now its junk Posted: Apr Tue 03, 2012 9:27 pm |
|
Joined: Dec Wed 05, 2007 11:08 pm Posts: 693 Location: Marietta/Moultrie GA
|
Tubenut wrote: I have 5 tube testers, never have I had to adjust the line Voltage control with a tube in. A number 27 tube's filament draw is moderate. For the reason above, I think this is a good practice. Especially if it's a special tube. Many old tube testers have a step control for the line adjust, IE: Heathkit and so on...... There is no need to be "jolting" the tube, when setting this control prior to the test is perfectly adequate.
Then, you've never had a Hickok TV7. It's right in the manual... to push the button, and adjust the line voltage control to bring the needle to the center mark, AFTER putting in the tube and turning on the tester. The readings are WORTHLESS, unless you do this... you have no idea how they compare to standard. Might as well not bother to calibrate the tester in the first place, unless you are willing to actually follow the instructions... That's how the US Army did it for MANY years... Regards, Gordon.
_________________ "It's the guys who think that attending meetings is real day's work that are the problem."- HepcatWilly (on AudioKarma)
|
|
| Top |
|
 |
|
Tubenut
|
Post subject: Re: now its junk Posted: Apr Tue 03, 2012 10:52 pm |
|
Joined: Oct Sun 11, 2009 10:06 am Posts: 1441 Location: British Columbia
|
|
Hey Gordon, No, I have not owned a TV-7. On most testers it's very easy to adjust the line Voltage without a tube installed. All the testers I have owned, there has never been a need to adjust this with a tube installed. My Heathkit IT-17 tells you to adjust the line Voltage first. This is the first manual I picked up. The needle stays well within the calibration markers when you plop a tube in. It would be interesting to know what kind of tube tester hcheetham has.
|
|
| Top |
|
 |
|
Stan Ski
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 12:13 am |
|
| Member |
 |
Joined: Jan Thu 01, 1970 1:00 am Posts: 7027 Location: Clinton Twp. Mi.
|
Tubenut wrote: Hey Gordon, No, I have not owned a TV-7. On most testers it's very easy to adjust the line Voltage without a tube installed. All the testers I have owned, there has never been a need to adjust this with a tube installed. My Heathkit IT-17 tells you to adjust the line Voltage first. This is the first manual I picked up. The needle stays well within the calibration markers when you plop a tube in. It would be interesting to know what kind of tube tester hcheetham has. This makes no sense at all. After making all the sure all the setting are correct i install the tube. Depending on how much current the tube is drawing you will see the meter move. How can it be possible to set the meter without any load on it? Have to ask you a question. What is with the font color? Do you like blue? Stan Ski
|
|
| Top |
|
 |
|
35Z5
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 1:15 am |
|
| Member |
 |
Joined: Jan Thu 01, 1970 1:00 am Posts: 8654 Location: Chesapeake VA
|
|
Many testers require the line be set after inserting the tube... It can be set prior but as mentioned when the tube is inserted there will be some voltage sag so setting will have to be increased... On my old Precision 920 I know approx how much increase is necessary, unless it's a battery operated tube, I set it slightly high and it usually only needs a minor tweaking...
My B&K707 automatically sets line, uses two #55 lamps to monitor line current similar to some Hickoks...
Tom
|
|
| Top |
|
 |
|
hcheetham
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 1:57 am |
|
Joined: Jan Thu 01, 1970 1:00 am Posts: 892 Location: Sodus,NY USA
|
|
I have a triplett tester.you have to bring it up to line then use other knob and turn left or right for the basic tests.today it seemed to do what it was supposed to.
_________________ Visit us on the web at: sunshineradiomuseum.org W2HJC
|
|
| Top |
|
 |
|
Tubenut
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 4:48 pm |
|
Joined: Oct Sun 11, 2009 10:06 am Posts: 1441 Location: British Columbia
|
Stan Ski wrote: Tubenut wrote: Hey Gordon, No, I have not owned a TV-7. On most testers it's very easy to adjust the line Voltage without a tube installed. All the testers I have owned, there has never been a need to adjust this with a tube installed. My Heathkit IT-17 tells you to adjust the line Voltage first. This is the first manual I picked up. The needle stays well within the calibration markers when you plop a tube in. It would be interesting to know what kind of tube tester hcheetham has. This makes no sense at all. After making all the sure all the setting are correct i install the tube. Depending on how much current the tube is drawing you will see the meter move. How can it be possible to set the meter without any load on it? Have to ask you a question. What is with the font color? Do you like blue? Stan Ski Wow... Hey Stan, It's in the IT-17 instructions. I find this topic really amusing. It's amazing how many people can't wrap their heads around this. I have never destroyed a single tube in a tube tester with this "way of doing things." It would be a shame to accidently turn the filament control up instead of adjusting the line control. As mentioned before, none of my testers "sag" to the point of needing re-calibration with the line Voltage. The bottom line is, If your tester requires re cal with every different tube you plug in, "do it!" Mine don't. My Mercury tester specifically tells you " not to re-adjust the needle." My BK 707, well 35Z5 explains it above....So on.
Yes Stan, I like Blue. It's nice to have a little color in "Pleasantville." You have seen the movie, haven't you? This site "echo's that movie Or how about Grumpy Old Men.
Hi hcheetham, It sounds like your tube may have had a weak filament or air in the bulb. I would be careful with the tester until you know for sure what the issue is.
|
|
| Top |
|
 |
|
easyrider8
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 8:10 pm |
|
Joined: Jan Thu 01, 1970 1:00 am Posts: 6768 Location: Minnesota
|
|
For the Heathkit IT-17 Step 1 says to adjust the line control without the tube in the socket. Step 7 says to readjust the line control after the tube is installed.
This is normal procedure for many tube testers.
Dave
|
|
| Top |
|
 |
|
Tubenut
|
Post subject: Re: now its junk Posted: Apr Wed 04, 2012 10:27 pm |
|
Joined: Oct Sun 11, 2009 10:06 am Posts: 1441 Location: British Columbia
|
easyrider8 wrote: For the Heathkit IT-17 Step 1 says to adjust the line control without the tube in the socket. Dave Exactly! Out of all the years I have owned my IT-17 I have never had to do step #7 But since everyone is so keen on moving that control with a tube installed, maybe I should pop a tube in it and spend some time aligning the line Voltage accuracy to .01PPT. I have close to 40000 tubes, I guess I should get started. I wonder if any of the members on here have hooked a Spectracom time base to their tube tester so they also have perfect 60.0000000000 Hz. You know that will affect the accuracy of the test. Maybe a line conditioner? This is a great thread!
|
|
| Top |
|
 |
|
vitanola
|
Post subject: Re: now its junk Posted: Apr Fri 06, 2012 4:09 am |
|
Joined: Apr Tue 03, 2007 1:31 am Posts: 3354 Location: Jonesville, MI
|
Tubenut wrote: easyrider8 wrote: For the Heathkit IT-17 Step 1 says to adjust the line control without the tube in the socket. Dave Exactly! Out of all the years I have owned my IT-17 I have never had to do step #7 But since everyone is so keen on moving that control with a tube installed, maybe I should pop a tube in it and spend some time aligning the line Voltage accuracy to .01PPT. I have close to 40000 tubes, I guess I should get started. I wonder if any of the members on here have hooked a Spectracom time base to their tube tester so they also have perfect 60.0000000000 Hz. You know that will affect the accuracy of the test. Maybe a line conditioner? This is a great thread! An IT-17 is an emissions tester, is it not? I wouldn't think that it would be as critical as a good Dynamic Mutual Conductance tester, like the TV-7 mentioned above, or my Hickock 539 or my trusty old I-177B. Emissions testers were the dope back in the early 'thirties, but by the time the War came just about every tech who could afford one jumped for the more sophisticated unit. A dynamic tester is also helpful when selling tubes. Back when I was flogging fifty to seventy high-end audio tubes a week I found it worthwhile to pick up a calibrated Hickock 600 (a less sophisticated contemporary of the 539) at some considerable expense simply because many of my prospective customers in the former Greater East Asia Co-Prosperity Sphere were familiar with that common unit. As for the line adjustment, well I've been adjusting the line adjustment whilst the power transformer is under heater load AS PER TESTER MANUFACTURER'S INSTRUCTIONS for nearly forty years, now, ever since I picked up my first surplus I-177B when I was in fourth grade. Never had a tube go out while under test due to incorrect line adjustment, either, even though I've always tended to test a lot of the more delicate stuff, '99's, WD-11's, 101F's, 216's etc. I would assume that as long as one's tester is reasonably well calibrated and one follows the manufacturer's instructions, and the tester os of reasonable quality one will get reasonably repeatable results. Of course, even the best tester offers only a general gauge of quality. many weak tubes perform well in certain applications, and some strong tubes don't perk very well in others. What sort of tubes do you deal with, by the way? Early stuff, or the late heater cathode things?
_________________ "Gentlemen, you have come sixty days too late. The Depression is over" Herbert Hoover, June 6, 1930
|
|