 Post subject: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 7:36 pm 
Member
User avatar

Joined: Dec Tue 22, 2020 4:54 am
Posts: 92
Location: Montreal (Québec) Canada
I just bought the Heathkit IM-32 vtvm in good condition.
But I wonder why they put 2 capacitor of 0.05uf and on the parts list of IM-11 or IM-28 the value is 0.005uf.
(I couldn't find the shematic diagram of IM-32, but I believe they are the same)

And I would like to know if you would change them. And there is also the capacitor of 0.047uf 1600v but it seems to be in very good condition I can maybe keep it?

(I did DC and AC tests, check resistance and everything seems to be fine

Thank you

Serge :)


Attachments:
20210928_131351_HDR.jpg
20210928_131351_HDR.jpg [ 680.33 KiB | Viewed 470 times ]
20210928_135532_HDR.jpg
20210928_135532_HDR.jpg [ 603.37 KiB | Viewed 470 times ]
Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 8:18 pm 
Member
User avatar

Joined: Jan Tue 31, 2012 1:55 am
Posts: 18306
Location: Mountains of Mourne. Ireland.
https://www.nostalgickitscentral.com/he ... a_im32.pdf


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 8:53 pm 
Member
User avatar

Joined: May Wed 18, 2011 2:40 am
Posts: 5749
Location: Littleton, MA
serge64 wrote:
But I wonder why they put 2 capacitor of 0.05uf and on the parts list of IM-11 or IM-28 the value is 0.005uf.

You are looking at the wrong capacitors in the IM-11 and IM-28. The 0.05 µF capacitors in your photo are in the AC measurement circuitry, around the 6AL5 tube.

In the IM-11 and IM-18, they are 0.02 µF.

In the IM-13 and IM-28, they are 0.05 µF.

The 0.005 µF capacitors are on the grids of the 12AU7. They are the same value in all of the IM-11, IM-13, IM-18, IM-28, and IM-32

_________________
Steve Byan https://www.byan-roper.org/steve/steve-at-play/


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 9:05 pm 
Member
User avatar

Joined: May Wed 18, 2011 2:40 am
Posts: 5749
Location: Littleton, MA
serge64 wrote:
And there is also the capacitor of 0.047uf 1600v but it seems to be in very good condition I can maybe keep it?

(I did DC and AC tests, check resistance and everything seems to be fine)

Did you perform a DC leakage test at 1600 volts?

What is your "AC test"?

You checked its resistance, but at what voltage? What voltage does your VOM, VTVM, or DMM apply? 1.5 Volts? 15 Volts?

The only way to be sure it is OK is to test it at full rated voltage.

That John E. Fast capacitor may be a paper-in-oil capacitor, you might be able to sell it as a "tone cap" to a guitarist for enough cash to buy a handful of replacements.

_________________
Steve Byan https://www.byan-roper.org/steve/steve-at-play/


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 10:30 pm 
Member
User avatar

Joined: May Wed 18, 2011 2:40 am
Posts: 5749
Location: Littleton, MA
serge64 wrote:
(I did DC and AC tests, check resistance and everything seems to be fine)

Do you mean you tested the capacitor, or that you tested the DC, AC, and resistance functions of the VTVM?

_________________
Steve Byan https://www.byan-roper.org/steve/steve-at-play/


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Tue 28, 2021 11:10 pm 
Member

Joined: Jun Fri 19, 2009 6:34 pm
Posts: 11008
Location: Long Island NY
Do not replace the 0.005 ceramic disc capacitors on the grids of the 12AU7. Though they are ordinary Z5U parts, they were carefully selected at the factory for minimum leakage. This is necessary to prevent drifting of the pointer. They are almost certainly still good, and if you were to replace them with other brands or types of capacitors you could end up chasing the meter pointer up and down the scale with the zero knob as the circuit drifts.

As for the other capacitors, the IM-32 was produced in 1962 and 1963. By then Heathkit had largely shifted away from paper capacitors to polyester (Mylar) film. So except for the 20-uF power supply electrolytic, which should be replaced if it hasn't been already, the caps that are in there might still be okay. The only way to know for sure would be to measure their leakage resistance. If either of the two 0.05-uF caps are leaky, the 6AL5 half voltage doubler circuit will not work properly and the AC voltage readings will be off.

As for the 0.047-uF, 1600 volt cap, its role is to block DC voltages from the AC measurements. You can test this by setting the meter to read AC volts, then connecting the AC probe to a high voltage DC source (with the ground clip connected). Set the voltage scale to a high voltage then work your way down. The needle may "kick" as the capacitor charges, but it should settle back to zero eventually. If you continue to get a reading then either the capacitor is leaky or the voltage source you are measuring has some AC in it.

_________________
"Hell, there are no rules here--we're trying to accomplish something!"

Thomas A. Edison


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 12:52 am 
Member
User avatar

Joined: Dec Tue 22, 2020 4:54 am
Posts: 92
Location: Montreal (Québec) Canada
For the capacitor 0.047uf.
I put 85v dc in AC max position. and I went down the AC voltage scale.
The needle does not move at all, except that if I go from 5v to 1.5v it moves a little

(The only capacitor I changed is the 20uf and the model 10 diode. I changed it to a 1n4001 diode.)


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 1:19 am 
Member
User avatar

Joined: Dec Tue 22, 2020 4:54 am
Posts: 92
Location: Montreal (Québec) Canada
I didn't have a dc probe so I took a Boonton probe that was not working and emptied it.
I put 2 resistors of 2M ohm 2 watts in parallel in a box connect to the jack

This is what i need to measure dc ?


Attachments:
20210928_200137.jpg
20210928_200137.jpg [ 716.68 KiB | Viewed 406 times ]
Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 1:21 am 
Member

Joined: Jun Fri 19, 2009 6:34 pm
Posts: 11008
Location: Long Island NY
If you disconnect the 85 volt supply and switch from the 5VAC to 1.5 VAC scale, does the zero change? On some meters it changes a little so you have to make sure zero is still zero.

_________________
"Hell, there are no rules here--we're trying to accomplish something!"

Thomas A. Edison


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 1:34 am 
Member
User avatar

Joined: Jan Tue 31, 2012 1:55 am
Posts: 18306
Location: Mountains of Mourne. Ireland.
For your viewing pleasure.
Manuel Caldeira. Madeira, Portugal.

How to Reform Electrolytic Filter Capacitors on tube radio or amp.
Image
Watch 22:50 https://www.youtube.com/watch?v=ifWfvaIbWZY


Capacitor leakage tester design & build - Part 1
Image
Watch 45:19 https://www.youtube.com/watch?v=0pySErvzUIY

Capacitor leakage tester design & build - Part 2
Image
Watch 18:20 https://www.youtube.com/watch?v=cW9hHjOxmQw

Capacitor leakage tester design & build - Part 3
Image
Watch 31:20 https://www.youtube.com/watch?v=NBUiaw18eOU

Capacitor leakage tester design & build - Part 4
Image
Watch 16:01 https://www.youtube.com/watch?v=X4B79BYIlbY


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 10:31 pm 
Member

Joined: Feb Fri 22, 2008 3:28 am
Posts: 606
You might want to read my comments on this thread in regard to the large thumbwheel knobs on that meter.

viewtopic.php?f=8&t=385450

Meter does work well; I like the large size meter movement like the one used in the IM-13 meter. The weak point is those knobs; I doubt they expected they would still be in use over 50 years later.


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Wed 29, 2021 10:46 pm 
Member
User avatar

Joined: May Wed 18, 2011 2:40 am
Posts: 5749
Location: Littleton, MA
serge64 wrote:
This is what i need to measure dc ?

It would be better to put the resistor in the probe. Its purpose is to isolate the capacitance of the shielded cable from the tip of the probe.

As you have it, it does not isolate the cable capacitance. You might as well have installed it inside the meter.

_________________
Steve Byan https://www.byan-roper.org/steve/steve-at-play/


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Thu 30, 2021 2:05 am 
Member

Joined: Jan Tue 10, 2012 8:39 am
Posts: 2815
I am trying to sort out the parameters for evaluating capacitor leakage. As I see it, it takes two forms.

One is coupling capacitors where leakage causes bias shifts in tube amplifiers. This needs calculation to decide how much leakage is allowable.

The other is power supply filtering, where leakage can raise capacitor temperature due to power dissipation, which could cause thermal runaway.

There are other situations as well, such as timing networks/RC oscillators.

Has anyone here worked out any permissible leakage currents? Or perhaps knows a reference that doesn't skim over the subject by suggesting rules of thumb. The definitive answer probably lies in the capacitor data sheet, assuming the circuit designer took the trouble to see what leakage his capacitor choice specified.


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Thu 30, 2021 3:08 am 
Member

Joined: Feb Fri 22, 2008 3:28 am
Posts: 606
Someone posted some tables for (allowable) electrolytic capacitor leakage on an older ARF thread; not sure where they were sourced. They were intended I think more to decide if the capacitor should be tossed or not.


Top
 Profile  
 
 Post subject: Re: heathkit vtvm model im-32
PostPosted: Sep Thu 30, 2021 3:30 am 
Member
User avatar

Joined: Dec Tue 22, 2020 4:54 am
Posts: 92
Location: Montreal (Québec) Canada
stevebyan wrote:
serge64 wrote:
This is what i need to measure dc ?

It would be better to put the resistor in the probe. Its purpose is to isolate the capacitance of the shielded cable from the tip of the probe.

As you have it, it does not isolate the cable capacitance. You might as well have installed it inside the meter.


I thought about doing this but I had only 2X 2M ohm 2 watt resistor in hand and it did not fit into the probe.

I'm going to buy a 2 watt 1M ohm resistor and put it inside the probe. Thank you.

And thank you all for helping me.

Serge :)