Update #3Retired Radio Man asked if I had checked the linearity of the meters before embarking on the restoration. I hadn’t, but fortunately hadn’t started replacing the bad resistors. It would have been disappointing to put all that work into a unit only to find that the meter movement was kaput. So, I went about checking out the meter movements.
My first task was to zero the movements.

RCA didn’t provide a zero-adjustment screw for the WV-98 meters. Instead they provided a small hole in the front panel, covered by a plug, that provides access to the zero adjustment tang on the meter movement.

You have to use a small tool such as a tiny screwdriver to reach in and nudge the adjustment tang. You also have to be very careful to not put the tool in too deep, as that could damage the meter armature.

As a final obstacle, the meters came from RCA with a piece of cellophane tape covering the hole in the meter bezel that gives access to the adjustment tang. You can try puncturing the tape, but if you do you run the risk of the tool going too deep and damaging the movement.
The tape had been removed from one of my meters, so that was easy to zero. On the other two, I had to remove the meter from the VTVM front panel. Then I discovered the first meter was out of balance, so I had to remove it from its front panel as well.

To remove the meter from the front panel, remove these four nuts and lockwashers. (The other two screws held the upper support bracket to the front panel.)

Then push the top of the meter out from the front panel. The clearance is tight on the two lower screws, but with a bit of jiggling they will come free.

I carefully removed the meter cover to get access to the weights to rebalance the meter. To remove the cover, carefully release the two tabs at the top. These tabs fit into two recesses at the top of the meter housing. You can use a 3/16” bladed screwdriver to gently pry the top of the cover away from the meter case to release a tab.

I now had access to the quad. The quad is the cross-shaped piece mounted to the meter pivot. The needle attaches to one arm of the quad. In the photo you can just see the tail weight on the arm of the quad opposite the needle. It’s the helical spring just above the solder joint for one end of the armature winding.
The balancing process is described well in several posts here on the Antique Radio Forum, so I’ll just provide links to them instead of repeating the process here:
The next step was to check the linearity of the movements, using my HP 6920B meter calibrator. I collected data on the current at each cardinal point on the meter scale. I started the run at zero and increased the current to reach each point, up to full-scale. The error is calculated as a percentage of the measured full-scale for that particular meter.

The error of the meter from the K114 run (red plot) was a bit high on the upper half of the scale, although the total variation is within ±1%. The meter from the L60 run (yellow plot) looks better. But when I ran the meter from the K117 run (blue plot), the error was very high at mid-point, and the meter seemed sluggish getting past the half-way mark. I ran a curve starting from full scale and going back down to zero, to see if there was a lot of hysteresis (green plot). Some hysteresis is evident, but it’s not very bad.

I suspected there was some foreign matter interfering with the movement of the armature, so I removed the meter card and went fishing with a sliver of Scotch tape in the gap. I didn’t find anything. so my next guess was some slight corrosion of the meter pivot. This VTVM showed evidence of battery leakage at some point in its life.

I remembered a
post by Chris H here on the Antique Radio Forum. He had suggested exercising a meter with a function generator to clear up minor corrosion in the bearings.
I set my B&K function generator to produce a 0.125 Hz triangle wave with a DC offset so the waveform went between zero and about 2 Volts. I connected each of the three meters to the function generator output through a 10K Ω resistor, giving a full scale of about 200 µA at the peak of the triangle wave.
Here's a video of the process:
https://www.byan-roper.org/steve/_Media/img_8730.m4vI exercised the meters for a few hours one evening. Afterward, the problematic meter ran smoothly through mid-scale.
I then took a series of measurements on each meter, both running up the scale and running back down. Again, the error is calculated as a percentage of the measured full-scale for that individual meter. I took four sets of measurements in each direction for the run K114 and run L60 meters, and two sets of measurements in each direction for the run K117 meter. I plotted the average of these sets of measurements, along with the initial, pre-exercise, measurements.

This is the plot for the run K114 meter. Comparing the linearity of the initial (red plot) and final (blue plot), you can see a small improvement. The hysteresis looks pretty bad though, nearly a half of a percent in the worst case.

Here is the plot for the run K117 meter. Exercising it with the function generator greatly reduced the sticking at the half-scale mark. The overall linearity is now pretty good.

This is the plot for the run L60 meter. The initial (red plot) and final (blue plot) have about the same total error (about 1.2%), but the post-exercise error is shifted up. There is also quite a bit of hysteresis evident.
L60 is a fairly late run, with the new-style RCA logo and the new blue paint scheme in place of the old gray hammertone. I suspect the meters in these later runs are a bit crappier as a result of cost-reduction.
Now that the meters checked out as usable, ...

… the next step was to proceed with replacing the out-of-spec resistors. Here I’ve completed reworking the range switch for the unit from run K114.

And here is the completed PCB for the unit from run K114.

The “before” photo for the K117 run.

All of the DC range resistors had to be replaced.

Another view of the K117 run range switch with new resistors.

And here’s the K117 run PCB. Sorry it’s out of focus.

Here’s the range switch from the run L60 unit. Only a few of the resistors needed to be replaced.

The run L60 PCB with its replacement resistors.

I reworked the power supplies by replacing the original electrolytic with a modern radial-leaded part, and the rectifier by a 1N4006.

I’ve reinstalled the range switch and am about to reconnect all the wires to the PCB.

Reconnecting the range switch wasn’t too bad. I just followed the kit instructions. It was very helpful to have recorded the wire colors for each step, back when I disassembled it.
I’m undecided whether I’ll complete the L60 unit before moving on to the others, or do it assembly-line style and reinstall the range switches in the other units before reconnecting the controls and connector on the lower support bracket.